Standardized proteomic Throughput ANalyzer
Know your instrument. Community reference ranges for mass spectrometer QC.
Expected performance ranges established by the community. Use the filters below to find your instrument and method. Ranges update automatically as more labs contribute.
Each point represents an anonymous submission. Distributions show the community range for each instrument family. Hover for details.
These are the metrics the 2024 proteomics QC literature (NIST MSQC, QCloud2, CPTAC, PTXQC) considers the real LC-health signals — they catch failures BEFORE identifications collapse. Unlike IPS (which is a depth rank), these don't depend on how many peptides you identified.
Your instrument's QC trend over time, with community reference bands (mean ± 1/2/3σ from all submissions of the same instrument family). Use this to answer: "Is my instrument drifting, and am I above or below the community baseline?"
All STAN benchmarking uses the
Pierce HeLa Protein Digest Standard
(Thermo Scientific, cat# 88328 / 88329). This is a tryptic digest of the
HeLa S3 cell line containing >15,000 proteins with <10% missed cleavages.
Using a single, commercially available standard ensures every lab starts
from the same sample. Differences in metrics reflect instrument and LC
performance, not sample preparation variability.
Buy 20 μg (88328) ·
Buy 5 x 20 μg (88329)
A composite score computed entirely from search output — no reference
run, no blanks, works from the very first QC injection.
DIA:
30% precursor depth ·
25% spectral quality (frags/precursor) ·
20% sampling (pts/peak) ·
15% quant coverage ·
10% digestion
DDA:
30% PSM depth ·
25% mass accuracy ·
20% sampling (pts/peak) ·
15% hyperscore ·
10% digestion
90-100 Excellent
80-89 Good
60-79 Marginal
<60 Investigate
Every community submission searches the exact same frozen, hash-verified
human UniProt FASTA and predicted spectral library. Identical upstream
parameters mean differences in output reflect instrument performance,
not search configuration.
Primary: Precursors (DIA) / PSMs (DDA) — purest instrument signal
Secondary: Peptides — slight sensitivity to search settings
Context: Proteins — shown for reference, not used for ranking (still affected by inference algorithm)
Health: IPS, missed cleavages, charge distribution
The number of MS2 scans sampling each precursor's elution profile directly
determines quantitation accuracy
(Matthews & Hayes, 1976).
12+ points: reliable quantitation
6-12 points: minimum for acceptable accuracy (depends on peak shape)
<6 points: systematic quantitation error increases rapidly
At high SPD with short columns, cycle time can exceed peak width.
These are guidelines — actual error depends on peak symmetry and
integration method. Track this metric to find the throughput limit
for your setup.
All submissions are anonymous by default. Click column headers to sort. Use the export button to download as CSV.